Posted: Sat Jan 05, 2019 7:28 pm Post subject: 100X objective tests |
|
|
Testing 100X objectives on a metallurgical scope.
First a baseline 40X stack
100X dry objective
And the big surprise - 100X oil objective
 _________________ Michael Reese Much FRMS EMS Bethlehem, Pennsylvania, USA |
|
Back to top |
|
 |
rjlittlefield Site Admin

Joined: 01 Aug 2006 Posts: 20177 Location: Richland, Washington State, USA
|
Posted: Sat Jan 05, 2019 7:38 pm Post subject: |
|
|
Is the oil objective used dry here?
--Rik |
|
Back to top |
|
 |
Lou Jost
Joined: 04 Sep 2015 Posts: 3840 Location: Ecuador
|
|
Back to top |
|
 |
Olympusman
Joined: 15 Jan 2012 Posts: 4271
|
Posted: Mon Jan 07, 2019 10:21 am Post subject: Objective test |
|
|
I should have put the specs of the objectives in the original post, which I will do from now on. The objectives are:
NIKON MPlan 40 0.65 210/0
The dry 100x objective may have come with an Amscope kit.
100/1.25
160/0.17
The 100X oil is a SWIFT objective that I have previously tested against my other oil objectives in transmitted light microscopy. This was the first time I tried it on a metallurgical scope with immersion oil and am very impressed by its performance.
SWIFT S-MICRO
100 1.25 Oil
160/0.17 _________________ Michael Reese Much FRMS EMS Bethlehem, Pennsylvania, USA |
|
Back to top |
|
 |
Pau Site Admin

Joined: 20 Jan 2010 Posts: 4859 Location: Valencia, Spain
|
Posted: Mon Jan 07, 2019 10:47 am Post subject: |
|
|
Quote: | The dry 100x objective may have come with an Amscope kit.
100/1.25
160/0.17 |
NA 1.25 is an oil immersion objective for sure. The maximum NA attainable in air is 0.95 (and with water immersion 1.2, I think).
Not surprising the bad results without oil _________________ Pau |
|