SEM Issue

A forum to ask questions, post setups, and generally discuss anything having to do with photomacrography and photomicroscopy.

Moderators: rjlittlefield, ChrisR, Chris S., Pau

Noah212
Posts: 30
Joined: Mon Sep 19, 2011 4:46 pm
Location: North East, U.S.A.
Contact:

SEM Issue

Post by Noah212 »

I had the incredible opportunity of using a scanning electron microscope today, although I encountered a significant issue and am hoping some of you here can offer some insight. The butterfly scales I imaged were visibly waving back and forth/flexing on the SEM's monitor even though they were in a vacuum chamber and on a vibration isolation table.

Here are the images:

(Besides converting from .tif to .jpg, these are straight from the SEM)

Image

Image

Image


The imaging lab where I worked today is on the fifth floor (though each floor is probably 3 stories) and a subway runs underneath. Could this be the source of the movement even though the SEM is on a vibration isolation table? Other specimens didn't suffer from this strange movement so I don't think that the subway and height actually affected anything.

My theory is that the scales, which are overlapping on each other and are really quite thin, don't have enough support and thus are somewhat free to move around, though I'm not quite sure what forces would cause this movement seeing as the specimen is in a vacuum chamber and the SEM itself is on a vibration isolation table.


Any ideas?

rjlittlefield
Site Admin
Posts: 23625
Joined: Tue Aug 01, 2006 8:34 am
Location: Richland, Washington State, USA
Contact:

Re: SEM Issue

Post by rjlittlefield »

Noah212 wrote:My theory is that the scales, which are overlapping on each other and are really quite thin, don't have enough support and thus are somewhat free to move around, though I'm not quite sure what forces would cause this movement seeing as the specimen is in a vacuum chamber and the SEM itself is on a vibration isolation table.

Any ideas?
I have no experience in this area. But it occurs to me that the electron beam itself will carry some momentum, and because the beam hits each scale only periodically, there's some opportunity to set up resonances.

Have you tried changing the beam intensity to see if the motion changes?

--Rik

Pau
Site Admin
Posts: 6071
Joined: Wed Jan 20, 2010 8:57 am
Location: Valencia, Spain

Post by Pau »

I agree with each Rik's word
Pau

dmillard
Posts: 639
Joined: Tue Oct 24, 2006 7:37 pm
Location: Austin, Texas

Post by dmillard »

It's been years since I have done any SEM (using Polaroid PN 55!), but perhaps the streaking and distortion may come from a charge build-up, and resulting movement, due to insufficient sputter coating on the non-conductive scales, or failure to adequately ground them on the stub on which they are mounted.

Regards,
David

g4lab
Posts: 1437
Joined: Fri May 23, 2008 11:07 am

Post by g4lab »

I also have not done SEM since the days of P/N 55

But I agree with the charge build up theory. The scales are mostly insulating even though you may have coated them.

There could also be thermal effects in the electron beam since the scales don't have much mass thermal nor otherwise. They are designed and intended to be light and fluffy.

flyer2o12
Posts: 223
Joined: Mon Jan 09, 2012 5:06 pm
Location: Ireland
Contact:

Post by flyer2o12 »

Hi,

I do SEM all the time and I have experienced this in my first few weeks of using it. The result is a charge build up in your sample which then deflects the electron beam. The reason for the charge accumulation is because there is insufficient amount of gold covering applied to your sample. So try to double or treble the sputtering time of your sample or you can experiment.