Greetings,
This is my first attempt with my new 100x objective Nikon LU Plan BD 100x NA 0.80.
Nikon D610 + Raynox DCR-150 + MJKZZ rail + 3xflash
Stack 64 frames, 0,9 um step size
Working distance is 3,5mm so it's pretty hard to bring light to subjects like eyes, but I'll do some more testing. Also the step size should be smaller but there are some mechanical limitations...
Vanessa atalanta white spot on the wing.
100:1
Crop 100:1
I'm happy that there is an obvious improvement in detail over 50:1 and NA 0.55. Also I like the 3d look of the ridges.
This is the same kind of wing with the same setup and Nikon LU Plan 50x NA0.55. Both pictures have some crop and processing, nothing too fancy.
50:1
Crop 50:1
The 50x can be better, it's kinda flat here because of the front-ish lighting.
50:1 and 100:1
Moderators: rjlittlefield, ChrisR, Chris S., Pau
This is quite an improvement!
With some levels adjustment and sharpening it looks really good.
With some levels adjustment and sharpening it looks really good.
It depends of the subject, exigence, illumination and maybe other factors...with a diatoms test stack I did while testing my newly acquired Wemacro coupled to the microscope I found that 0.35um was the best step with a 25/0.65 objective, much smaller than the generally accepted (and well proven) standardsSaul wrote:0.9um step worked pretty good, I'm using 0.5um for 50x. Maybe overkill ?
Pau
-
- Posts: 1965
- Joined: Sat Oct 07, 2006 10:16 am
- Location: Bigfork, Montana
- Contact:
- rjlittlefield
- Site Admin
- Posts: 23604
- Joined: Tue Aug 01, 2006 8:34 am
- Location: Richland, Washington State, USA
- Contact:
I am no longer surprised by such reports. At large NA, proper step size seems to vary a lot depending on the subject and illumination. A lot of this is due to utilized aperture effects that make the subject appear to wiggle around laterally, in addition to going into and out of focus. But sometimes I wonder if there is also some important structure in the lightfield reflected/diffracted off a 3D surface that is not captured by the utilized aperture concept either. The standard formula for 1/4-lambda wavefront error (=lambda/(NA^2)) is derived from a planar model and happily ignores all such effects, which means that it's sort of a "necessary but not sufficient" condition.Pau wrote:It depends of the subject, exigence, illumination and maybe other factors...with a diatoms test stack I did while testing my newly acquired Wemacro coupled to the microscope I found that 0.35um was the best step with a 25/0.65 objective, much smaller than the generally accepted (and well proven) standardsSaul wrote:0.9um step worked pretty good, I'm using 0.5um for 50x. Maybe overkill ?
--Rik
Hello
Nice pictures and very interesting!
There is a relation between Apertur and depth of field (DOF).
DOF (mm) = Lambda (mm) / (NA x NA)
In reality, i do not know how high the aperture exactly is!
The resolution of a bad objektiv is low, and the depth of field is higher.
There is a relation between Apertur, quality and resolution.
Everything is theoretical.
In practice, I can easily measure the resolution.
So I'm counting from an actually measured resolution to the aperture.
Apertur = (LP/mm x lambda) / 2
(1000LP/mm x 0.00055mm) / 2 = 0.275 NA
and from this aperture to depth of field.
DOF (mm) = lambda (mm) / (NA x NA)
0.00055mm / (0.275 x 0.275) = 0.00727mm DOF
Directly from resolution to depth of field
DOF (mm) = lambda (mm) / (((LP/mm x lambda (mm)) / 2) x ((LP/mm x lambda (mm) / 2))
0.00055 / (((1000 x 0.00055) / 2) x ((1000 x 0.00055) / 2)) = 0.00727mm DOF
The basis of this calculation is the real measured resolution and the result is a depth of field close to the reality
The macro rail is not absolutely exact and sometimes a picture fails.
That's why (for focus stacking) I make 3 pictures within this depth of field.
This manner of calculation works well from 1:1 to 100:1.
Kurt
Nice pictures and very interesting!
There is a relation between Apertur and depth of field (DOF).
DOF (mm) = Lambda (mm) / (NA x NA)
In reality, i do not know how high the aperture exactly is!
The resolution of a bad objektiv is low, and the depth of field is higher.
There is a relation between Apertur, quality and resolution.
Everything is theoretical.
In practice, I can easily measure the resolution.
So I'm counting from an actually measured resolution to the aperture.
Apertur = (LP/mm x lambda) / 2
(1000LP/mm x 0.00055mm) / 2 = 0.275 NA
and from this aperture to depth of field.
DOF (mm) = lambda (mm) / (NA x NA)
0.00055mm / (0.275 x 0.275) = 0.00727mm DOF
Directly from resolution to depth of field
DOF (mm) = lambda (mm) / (((LP/mm x lambda (mm)) / 2) x ((LP/mm x lambda (mm) / 2))
0.00055 / (((1000 x 0.00055) / 2) x ((1000 x 0.00055) / 2)) = 0.00727mm DOF
The basis of this calculation is the real measured resolution and the result is a depth of field close to the reality
The macro rail is not absolutely exact and sometimes a picture fails.
That's why (for focus stacking) I make 3 pictures within this depth of field.
This manner of calculation works well from 1:1 to 100:1.
Kurt